squarepush3r
Member
Hey habeeb,
The probe is fine. There are no special requirements for storage of the probe according to the manual. The meter allows you to check the electrode properties for diagnostic purposes. The slope factor, % saturation offset setting, 100% saturation mV value and the 0% mV saturation value were all fine according to the Oakton tech I spoke with.
Once I received the meter, I reset the factory values and cleared the calibration data.
I've replaced the membrane on the electrode and refilled the electrolyte in the probe. I've also calibrated it before each session first setting the atm pressure to what wunderground.com says it is at the weather station closest to my house. It then gets calibrated to 100% satuartion in air, and then using a zero oxygen solution. I also adjust the salinity setting based on the concentration of the nutes using my EC meter.
About the only thing I haven't done to make sure readings are accurate is compare the readings to another DO meter.
Sam,
I like that idea better than H2O2. K+
Squarepush,
Nice idea. I'm just not willing to stress my gal for the sake of data.
critical DO level is 2-3, you could even bring it to 4 or 5 temporarily on the safe side to run a mini test.
-S