http://www.101science.com/chemistrypage2.htm said:You can make your own pH calibration/buffer solutions as follows.
* pH 4.00: Add 2.0 mL of 0.1 M HCl to 1000 mL of 0.1 M potassium hydrogen phthalate. (pH 4.00 is used for probe storage)
* pH 7.00: Add 582 mL of 0.1 M NaOH to 1000 mL of 0.1 M potassium dihydrogen phosphate.
* pH 10.00: Add 214 mL of 0.1 M NaOH to 1000 mL of 0.05 M sodium bicarbonate.
For short-term storage (up to 24 hours) place the electrode in pH-4 or pH-7 buffer solution. For long-term storage (more than 24 hours) Store the electrode in a buffer pH-4/KCl storage solution. You can prepare a storage solution by adding 10 g of solid potassium chloride (KCl) to 100 mL of buffer pH-4 solution. By storing the electrode in this solution, the reference portion of the electrode is kept moist. Keeping the reference junction moist adds to electrode life and retains electrode response time when the unit is placed back into use. If the electrode is inadvertently stored dry (NOT RECOMMENDED!), soak the probe in a pH-4 or pH-7 solution for a minimum of eight hours prior to service. The pH Electrode can be cleaned with a light dishwashing detergent solution. For heavy deposits the electrode may be carefully cleaned with a 0.1 molar solution of HCl for no more than 30 seconds, then washed in a light detergent solution. Immediately after cleaning, rinse the electrode in pure distilled water. If the electrode still fails to calibrate properly, or responds slowly, electrode replacement is recommended. Mold growth in the buffer/KCl storage solution can be prevented by adding a commercial growth inhibitor. This mold will not harm the electrode and can easily be removed using a light detergent solution. Do not use the electrode in solutions containing perchlorate, silver, or sulfide ions. Do not use it in hydrofluoric acid or in acid or base solutions with a concentration greater than 1.0 molar. The electrode may be used to measure the pH of sodium hydroxide solutions with a concentration near 1.0 molar, but should not be left in this concentration of sodium hydroxide for periods longer than 5 minutes. Using or storing the electrode at very high temperatures or very low temperatures (near 0°C) can damage it beyond repair.