I am looking for a good working theory on what could be causing a consistent history of the following discrepancy between these two models of meters.
Pic is showing the issue. Ninety-nine percent of the time the new & improved meter is reporting a measurement that is a full 0.25 or more too high. Sometimes higher like 0.40 or more.
New meter has a second new probe. Same thing happened on the probe that came in the box.
The two old meters tend to agree. Old meters have attached, one very new probe; and one not new probe.
It's frustrating that I cannot get consensus among these meters when are all are calibrated with the same Hanna brand solution. The new one has a two-point calibration with Hanna 4 and Hanna 7. All meters will indicate 7 when in the calibration solution. But when they go into the solution that I need to measure, I get this anomaly.
Maybe once in 9-12 months had I ever seen them all agree. It was like seeing a ghost. haha.
I don't really have time to run an analysis of the pH reading across a range of EC points although that would be interesting to see. This pic is a reading of nearly RO water.
My tendency is to think the problem is the new meter.
After I check the calibration, I forget that I thought that until I compare meters again.
Homage to AC/DC's "For those about to rock!"
Pic is showing the issue. Ninety-nine percent of the time the new & improved meter is reporting a measurement that is a full 0.25 or more too high. Sometimes higher like 0.40 or more.
New meter has a second new probe. Same thing happened on the probe that came in the box.
The two old meters tend to agree. Old meters have attached, one very new probe; and one not new probe.
It's frustrating that I cannot get consensus among these meters when are all are calibrated with the same Hanna brand solution. The new one has a two-point calibration with Hanna 4 and Hanna 7. All meters will indicate 7 when in the calibration solution. But when they go into the solution that I need to measure, I get this anomaly.
Maybe once in 9-12 months had I ever seen them all agree. It was like seeing a ghost. haha.
I don't really have time to run an analysis of the pH reading across a range of EC points although that would be interesting to see. This pic is a reading of nearly RO water.
My tendency is to think the problem is the new meter.
After I check the calibration, I forget that I thought that until I compare meters again.